Evaluation of the GlucoWatch biographer: an automatic and non-invasive glucose monitor

Janet Tamada¹, Lois Jovanovic², Satish Garg³, Neil Ackerman¹, Steven Fermi¹, Russell Potts¹, Michael Tierne¹ ¹Cygnus, Inc., Redwood City, CA, USA; ²Sansum Medical Research Institute, Santa Barbara, CA, USA; ³Barbara Davis Center for Childhood Diseases, University of Colorado Health Sciences Center,  Denver, CO, USA

Background
The GlucoWatch® biographer is designed to facilitate the acceptance and success of intensive diabetes management by providing frequent, automatic glucose readings. The GlucoWatch biographer is worn on the arm like a watch. Glucose is sampled through skin by application of a low-level electric current, a process known as reverse iontophoresis, and the extracted glucose is measured with an electrochemical biosensor. The biographer displays up to three readings per hour for as long as 12 hours, following a single glucose measurement taken with a traditional blood glucose meter three hours after application. The biographer features user-adjustable high and low alert settings to warn of potential hypo- or hyperglycemia and stores 4000 readings, which can provide long-term glycemic pattern data. Data integrity checks screen for potentially erroneous data and the biographer displays a “skip” message to the user instead of a measurement.

Objective
The objective of this clinical study was to assess the performance of the GlucoWatch biographer in a demographically diverse diabetic population in a variety of environmental settings.

Methods
The accuracy of the biographer vs serial blood glucose measurements was studied in diabetic subjects in a controlled clinic environment, a simulated home environment, and the home environment. The subject population consisted of over 400 adult subjects diagnosed with Type 1 or Type 2 diabetes who required insulin treatment. Performance was assessed by analyzing paired glucose readings between the biographer and a traditional blood glucose meter. A number of blood glucose meters for both calibration and comparative blood measurements were utilized. Over 1000 biographer uses and 14,000 paired glucose measurements were studied.

Results
Time profiles over 12 hours of measurement show close tracking of the biographer results with fingerstick blood glucose measurements. The figure  shows a sample plot. The mean difference (MD) between the two measurements is less than 5 mg/dl over all studies. The mean absolute value of the relative difference (MARD) is approximately 20%. Over 94% of the data was in the clinically acceptable Error Grid A and B zones. Overall performance was comparable in all environments as shown in the Table.

Conclusions
The GlucoWatch biographer provides frequent measurements of glucose over a 12-hour period with good clinical accuracy. It can detect trends and track patterns in glucose levels in diabetic subjects with diverse demographic characteristics. The GlucoWatch biographer should facilitate improved diabetes management by the diabetic patient and caregiver.